Fixing a laboratory waterbath
  • eccerr0r
    Solder Sloth
    • Nov 2012
    • 8716
    Re: Fixing a laboratory waterbath
    The reasons why I'm doubting relay:
    - Voltage of the relays are typically printed on the relay
    - Switch characteristics are typically printed on the relay
    - It appears to be controlling the triac if that really is a triac there - versus a load. Again I don't see the connections so have to make assumptions with the incomplete data here.
    - tracing the input path, what a mess, the device is being driven with very little current, not enough for a relay due to a resistor in series with the current path.
    My current guess: "optoisolator" in quotes as a guess someone potted a LED and some optical sensors in that plastic.
    But this is still just a guess. But if that device feels like it has metal in it then fine it can be a relay and will need to retrace the circuit more carefully. Really would like a picture with all the boards assembled together to make sure things are the way I thought they are connected.
    • vrasp
      • Nov 2019
      • 194
      Originally posted by R_J
      You need to post a clear (in focus) high res, picture of the trace side of the board the relay is mounted on.
      Looking at the top suggests the small bridge rectifier and the 470µf capacitor are the supply for the relay, the to220 transistor controls the relay coil, The small glass diode next to the relay will be across the relay coil. I suspect the relay coil is either 12 or 24v based on the voltages marked on the transformer
      This is a picture with diagram from the very beginning of this "journey". Someone from reddit made this for me. I think it s useful in this case. You can see the connections where the relay is, if it really is one. I think it is though, but that's my non professional opinion.
      It would make sense that it is. The lower current side is connected to the pins (here labelled heater control signal) that go to the control PCB (not seen here) and activates the high current side which is connected to the gate. When the relay closes the circuit, the gate recieves a current and the triac conducts current to the heating element. Again, thats my non professional understanding of this circuit.
      Edit: there is a component (looks like a zener but I ll have to check) that is not seen on this picture. It s above the red capacitor and in front of the relay. It connects the relay to the gate of the triac. Sorry if this is confusing. I m gonna take better pictures and get back to you.
      • vrasp
        • Nov 2019
        • 194
        It is really messy. It doesn't help that I replaced a bunch of stuff and sometimes on the other side of the boards to make it work with the space I have. I ve also blown a few things because of things touching the grounded case again due to lack of space. But these things were replaced and it has worked fine since. Until this recent issue.
        I ll try to tske some more pictures with more details.
        The triac is a NTE56010.
        • R_J
          • Jun 2012
          • 9635
          Now that I see the under side of the board, that to220 is likely a +5volt regulator ic and not the relay driver. The relay is driven by the heater control board so that would need to be traced to determine the relay coil voltage.
          • vrasp
            • Nov 2019
            • 194
            Here are some more pictures and a video. Hopefully, that helps.
            The red capacitor and the "relay", or whatever it is, have been removed.
            https://vimeo.com/575955271
            The burned trace next to the triac is from the previous triac. This one is new. There are some cables hanging that I used to make it easier to measure pins I couldn't get to. These will be removed once I put everything back together.
            • eccerr0r
              Solder Sloth
              • Nov 2012
              • 8716
              oh gawd now I see that the two boards go together component side to component side, I thought they were stacked... this makes a huge difference, and previous assumptions made earlier on may need to be invalidated... ugh...
              1. How did you get away without an insulator on the TRIAC? Was this repaired before you worked on it? If so I'm very sorry. If you did replace it, you missed an insulator.
              2. It looks like the 5V from the transformer board is actually the LED power supply after all, and the transformer may be connected wrong.
              3. What gave you the impression the relay or whatever it is ... is bad? Also relays are denser than optoisolators, does it actually feel dense? Does a magnet stick to it?
              Sigh. I hate remote debug. About ready to give up on this.
              • vrasp
                • Nov 2019
                • 194
                Oh man. I had no idea I needed an insulator (The previous triac had what looks like an insulator already on it). Maybe that explains the short. The triac did come with a sticky plastic part which must be it. I guess I might have to replace it again.
                Everything was working fine with the transformer set up that way. The 5v from the 7805 do go to the LED.
                The relay is somewhat heavy. I ll try the magnet test. I thought it was bad because the high current side is always closed and so has continuity. Also the trace that blew was where the relay was (you can see on the photo I put some copper wire with solder on it).
                (Please dont give up haha)
                • eccerr0r
                  Solder Sloth
                  • Nov 2012
                  • 8716
                  That trace blew because you didn't put the insulator on. You need both the kapton/mica/... insulator with heatsink compound, plus the plastic shoulder washer. If the old screw doesn't fit with the shoulder washer, you need to find a thinner screw, usually #4 or M2.5 screws are thin enough.
                  Since you blew that trace, yes you may have blew the (relay). You might be able to get away with an arbitrary 12V relay but I still think it's optical because relays are not fast enough.
                  Then again perhaps the original designer was an idiot and designed it this way... *facepalm*
                  BTW, the "input" side should be quite conductive if it's a relay, this would be a good test and shouldn't get damaged by the shorting TRIAC. If it were a true optoisolator it needs to be an LED input, and these would be conductive in only one direction... again if it really is a relay, this is a very poor design.
                  Oh and another thought... it could be a transformer. Now discerning a damaged transformer and a relay with simple tools...
                  • vrasp
                    • Nov 2019
                    • 194
                    What about the triac? Can I assume that since the trace next to the "relay" (or let's call the the black thing) blew, the triac didn't suffer any damage. Do you think I should replace it?
                    Both sides of the black thing have continuity (between pins of the same side) in both directions. A fridge magnet sticks to both sides shown on the pictures but not the other 2 sides.
                    The writing says:
                    1/0,66
                    01/L06
                    I m trying to think what other info I could provide. If the signal that comes from the green/blue PCB is a DC signal then maybe we can rule out a transformer since they work with AC only.
                    • eccerr0r
                      Solder Sloth
                      • Nov 2012
                      • 8716
                      Aha... yeah that looks like transformer, that bottom side shows a lot. It's probably is fine, all the better - finding a replacement of this one would be tough. If your triac doesn't appear shorted it may still be fine, else you may need to replace it again.
                      • R_J
                        • Jun 2012
                        • 9635
                        I was also thinking about this and I suspect it is a pulse transformer and not a relay. Likely similar to a Schurter IL series
                        What is your plan for the transformer with the leads ripped out?
                        • vrasp
                          • Nov 2019
                          • 194
                          That's great! Very helpful. Do you know why they put s transformer there? What is its purpose?
                          Thank you!
                          • R_J
                            • Jun 2012
                            • 9635
                            The transformer may step up the pulse from the drive circuit, but it could also be for isolation, isolating the driver board from the triac which would be connected to the ac line.
                            • eccerr0r
                              Solder Sloth
                              • Nov 2012
                              • 8716
                              yep same reason for putting a relay or optoisolator there, transformers also provide isolation. Of the three, relay is the dumbest (why have a triac then?) Transformer is "ok" but trickier to use, and optoisolator would be ideal.
                              Now the question is whether you ever heard it "click" ... a transformer making clicking sounds is not normal.
                              • vrasp
                                • Nov 2019
                                • 194
                                I never thought of a transformer that way..to isolate circuit. I always just saw it as voltage step up / step down. That s so useful.
                                No I never heard a click. I tried to listen when I was testing it. Nothing. I believe you guys are right and it s a transformer.
                                Thinking further about this, if you have a short on one side which drives up the current, the voltage will go up too and have repercussions on the other side of the transformer. So the circuits are physically isolated but still reactive to one another/energically interconnected. What benefit does the physical isolation provide?
                                • eccerr0r
                                  Solder Sloth
                                  • Nov 2012
                                  • 8716
                                  Isolation is to make sure two circuits cannot interact with each other with a single additional connection. They are mainly for making two circuits that can be connected together without unexpected interactions, and the more important reason: protection from electrocuting people using the circuit.
                                  • vrasp
                                    • Nov 2019
                                    • 194
                                    The ripped out leads are from the previous transformer that has been replaced since.
                                    Should the thermal paste be between the triac and mica or mica and heatsink? or both? Also, what is the triac metal back plate connected to inside the triac? I thought the issue was that the triac metal part was touching the heatsink and was therefore grounded, but I dont get continuity between that metal part and the heatsink.
                                    EDIT:MT1, MT2, and the gate all have continuity with eachother (even when the triac is not on the heatsink) which I think means it's damaged.
                                    • eccerr0r
                                      Solder Sloth
                                      • Nov 2012
                                      • 8716
                                      Thermal paste (use VERY little just like on a CPU) should be on all surfaces that sit between a heat source and the sink where it could have touched air.... so both sides of a mica or kapton insulator.
                                      If you have a silicone insulator, which feels rubbery, you usually don't need the paste - only the hard stuff.
                                      In a properly designed circuit especially with TRIACs you may not actually need the paste as the dissipation shouldn't be as high as semiconductors running half on or at high frequencies.
                                      The standard TO-220, the metal tab is connected to the center terminal of the three pins in general. If your center pin is connected to hot on your AC input... BANG dead short to ground.
                                      However if you dirty the contact from the tab to the heatsink you might get high resistance, though you don't want to leave that to chance - use the insulator and the shoulder washer if it was insulated before.
                                      There are all plastic coated TO-220FP that are isolated from the center pin.
                                      • vrasp
                                        • Nov 2019
                                        • 194
                                        Got it. Thank you!
                                        Just got this stuff in the photo. Hopefully I m all set. I got a capacitor to replace the red one in case it broke when thebshort happened. I have to wait until next week to be in the lab.
                                        I get an extra insulator on top of the one provided with the triac. Which one should I use?
                                        Any tips on how to solder the triac pins to a wire? I just put flux and solder on both ends and melt it together for a few seconds then add a few layers of shrink tube (Like in the previous photo of the triac). Maybe there is a "cleaner" way to do it - like some kind of specialty connectors I can get with a female end to insert the pins.
                                        • eccerr0r
                                          Solder Sloth
                                          • Nov 2012
                                          • 8716
                                          capacitor shouldn't be harmed unless the desolder job broke it.
                                          You should use one or the other insulator, just not both.
                                          Make sure you use that shoulder washer that came with the TRIAC. The screw must not touch the TRIAC either.
                                          Without "the helping hands" tool, probably just mount the TRIAC and then solder to the pins. Careful not to pull on the wire after it's been soldered. Should make it a bit easier to solder. Definitely tin the wires first as you were doing.
                                          You don't want a socket here.
                                          • learnmore
                                            Help choosing replacement transformer for Levoit 600S SMPS
                                            by learnmore
                                            Hello Community,
                                            I'm attempting to repair the SMPS on a Levoit 600S air purifier. The transformer is likely open. It has six pins. 2 pairs on the primary side (one pair tested open) and 1pair of legs on the secondary side. Legs on the secondary side are spread wide apart (end-to-end). I've circled them in blue. Dimentons of the transformer are 22x23mm. Marking on the transformer are EE16 1.9mH-B. Fuse for the powerboard is fuse 250V 3.15A. I've been unable to find an exact match and I'm hoping for some help here to find a suitable replacment.
                                            Peace
                                            LM...
                                            11-04-2025, 09:35 AM
                                          • ohren
                                            Original Xbox 1.0 PSU: Foxlink FTPS-0002 Rev. B. — 12 V low, 5 V high — transformer?
                                            by ohren
                                            Hello!
                                            I'm looking at an original Xbox PSU: Foxlink ftps-0002 rev. B. The 12 V output is too low, and the 5 V output is too (?) high. Unfortunately I don't have an oscilloscope but only a DMM for diagnosis.
                                            Measured voltages with no load:
                                            After rectifier diodes
                                            12V: 6.17 V
                                            5V: 5.67 V
                                            Secondary transformer pins AC measurement (really don't know if this says anything)
                                            12V: 2-2.5 V
                                            5V: 1.04 V
                                            I also have another (working) Xbox psu from Delta. The same measures there, in order, being: 10.79,...
                                            03-18-2023, 10:33 AM
                                          • mikey5791
                                            Is it possible for transformer substitution?
                                            by mikey5791
                                            Got this local made DC power supply (12/14V) given free as the transformer has melted. There is no marking or indication to identify what type or rating of transformer used.
                                            Fyi,. I had a busted autogate mainboard with transformer with marking 13-0-13 . Is this the 13v dc type transformer?
                                            My questions are
                                            1. How do i test to see if the transformer is in good working condition?
                                            2. Can this transformer (marking 13-0-13) be used to replace the melted transformer on the DC power supply?
                                            Hope you guys understand what i was asking. My intention is...
                                            11-03-2022, 02:11 AM
                                          • dennis_elx33
                                            3PHASE STEP-DOWN Transformer in Check
                                            by dennis_elx33
                                            Hello All,
                                            i would like to post an interesting topic regarding 3-phase stepdown transformer which we suspect to cause an overcurrent (138A) which resulted to power tripped off of the HV power supply (80KV output).
                                            Problem statement: Power tripped Off
                                            Initial Problem: Found 2 IGBT and 2 rectifier diodes shorted (burnt).
                                            Action Done:
                                            1) Replaced the faulty parts with good spare.
                                            2) Measured impedance of the 3phase input lines and found that L2 to L3 has 23.3ohm while other pairs are 51Kohm. This pins will...
                                            12-02-2024, 06:52 AM
                                            ASUS Transformer T101HA - X5DHD Hybrid (2-in-1) Book Specification for Upgrade or Repair
                                            This specification for the ASUS Transformer T101HA - X5DHD Hybrid (2-in-1) can be useful for upgrading or repairing a laptop that is not working. As a community we are working through our specifications to add valuable data like the T101HA - X5DHD boardview and T101HA - X5DHD schematic. Our users have donated over 1 million documents which are being added to the site. This page will be updated soon with additional information. Alternatively you can request additional help from our users directly on the relevant badcaps forum. Please note that we offer no warranties that any specification, datasheet,...