#261 |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() |
![]() |
![]() |
![]() |
#262 | |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() Quote:
I have not wired anything differently since then. The only difference is the TRIAC which is a NTE56010 (800v 15A https://www.farnell.com/datasheets/202004.pdf). Below is the schematic in the manual of the waterbath. I posted it earlier too. It's a general schematic for 220v and doesn't go into the details. Last edited by vrasp; 07-25-2021 at 08:57 AM.. |
|
![]() |
![]() |
![]() |
#263 |
Badcaps Veteran
Join Date: Jul 2011
City & State: Some times Sunny Jacksonville FL
My Country: USA
Line Voltage: 120 Volts 60 HZ
I'm a: Knowledge Seeker
Posts: 3,554
|
![]() The description of the device will tell if it is a micro controller or not
Now it could be done with ic chip like a LM339 or something similar where you have a reference value and compare to the input value V1 is controlled by something the question is what this is controlled by this is where you are going to have to trace where these connections go to it has to go to some type of ic chip to control the triac unless it to some type of resistance pot which I doubt very seriously that it was done this way One note when you use a device that mount to a heat sink and the heat sink is mounted to the circuit board and has soldering joints that are connected to something else you need to make sure that the tab does NOT have connection to the heat sink so you need use an ohm meter to check this There is a difference between the two devices one has the metal mounting tab and the other one dose not and if you did not mount the device correctly this might explain why the new device failure happened Now there are devices that have the metal tab that has no connection to the device but you must check them to be sure that this is the case the data sheet should tell you this information The one thing that this data sheet does not tell you if the metal tab has a connection to the device so have to go with the assumption that it has a connection unless you have tested this with an ohm meter
__________________
9 PC LCD Monitor 6 LCD Flat Screen TV 30 Desk Top Switching Power Supply 10 Battery Charger Switching Power Supply for Power Tool 6 18v Lithium Battery Power Boards for Tool Battery Packs 1 XBox 360 Switching Power Supply and M Board 25 Servo Drives 220/460 3 Phase 6 De-soldering Station Switching Power Supply 1 Power Supply 1 Dell Mother Board 15 Computer Power Supply 1 HP Printer Supply & Control Board * lighting finished it * These two repairs where found with a ESR meter...> Temp at 50*F then at 90*F the ESR reading more than 10% 1 Over Head Crane Current Sensing Board ( VFD Failure Five Years Later ) 2 Hem Saw Computer Stack Board All of these had ![]() All of the mosfet that are taken out by bad caps ![]() Last edited by sam_sam_sam; 07-25-2021 at 09:16 AM.. |
![]() |
![]() |
![]() |
#264 | |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() Quote:
There is an LM324 on the control board. Last edited by vrasp; 07-25-2021 at 09:42 AM.. |
|
![]() |
![]() |
![]() |
#265 |
Badcaps Veteran
Join Date: Jul 2011
City & State: Some times Sunny Jacksonville FL
My Country: USA
Line Voltage: 120 Volts 60 HZ
I'm a: Knowledge Seeker
Posts: 3,554
|
![]() |
![]() |
![]() |
![]() |
#266 |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() Ok so this is my new plan:
1) Remove the transformer. 2) Connect a 0.1A fuse on GATE and 15A fuse on MT1. 3) Connect a 100w bulb instead of the heating element. 4) Turn on the waterbath. Without the transformer, the TRIAC should be OFF. If a fuse of GATE or MT1 blows, the TRIAC must be getting shorted on the heatsink. 5) Short MT1 and MT2 to light up the bulb. GATE should not be getting any current from MT1 so if the GATE fuse blows there is something wrong there. (is that a relevant test to do?) 6) If the bulb lights up and all the fuses are fine, connect the transformer back onto the PCB and turn on the waterbath. If the GATE fuse blows, there is a fault on the input side of the transformer, so that's coming from the control board. If everything is fine... well then everything should be fine. Last edited by vrasp; 07-25-2021 at 09:42 AM.. |
![]() |
![]() |
![]() |
#267 |
Badcaps Veteran
Join Date: Jul 2011
City & State: Some times Sunny Jacksonville FL
My Country: USA
Line Voltage: 120 Volts 60 HZ
I'm a: Knowledge Seeker
Posts: 3,554
|
![]() Vwrasp
WHY are trying to be dangerous in what you are talking about doing ( steps 4 and 5 and 6 ) All you need to do is put one led of volt meter to the heat sink and and check for resistance on the three pins on the highest setting if you have any kind of resistance you have an issue you which you need to fix Then if this part is correct then undo the heater element wires and hook an incandescent light bulb and see if you are getting the signal to the transformer to turn ON the triac if not then you need figure out why this is not working correctly Last edited by sam_sam_sam; 07-25-2021 at 12:58 PM.. |
![]() |
![]() |
![]() |
#268 | |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() Quote:
I had checked for continuity between the pins and the heatsink and there was none. However, I did not try with the voltmeter on the max Ohm setting. I ll do that this time and also have a fuse just in case. I ve gone through enough TRIACs. |
|
![]() |
![]() |
![]() |
#269 | |
Badcaps Veteran
Join Date: Jul 2011
City & State: Some times Sunny Jacksonville FL
My Country: USA
Line Voltage: 120 Volts 60 HZ
I'm a: Knowledge Seeker
Posts: 3,554
|
![]() Quote:
|
|
![]() |
![]() |
![]() |
#270 | |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() Quote:
|
|
![]() |
![]() |
![]() |
#271 |
Badcaps Veteran
Join Date: Jul 2011
City & State: Some times Sunny Jacksonville FL
My Country: USA
Line Voltage: 120 Volts 60 HZ
I'm a: Knowledge Seeker
Posts: 3,554
|
![]() |
![]() |
![]() |
![]() |
#272 |
Solder Sloth
Join Date: Nov 2012
City & State: CO
My Country: USA
Line Voltage: 120VAC 60Hz
I'm a: Hobbyist Tech
Posts: 7,161
|
![]() what i'd like to see is how the wires connect from the PCB to the TRIAC... the wire colors don't seem to make sense from the limited photos of the wires in the existing pictures, alas it looks like a lot of splices are in place so things may not be as they seem.
|
![]() |
![]() |
![]() |
#273 |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() This is an old picture but the best one I have. It s the first TRIAC.
Blue: GATE Black: MT2 Orange: MT1 . I wired the subsequent ones similarly. Last edited by vrasp; 07-25-2021 at 09:33 PM.. |
![]() |
![]() |
![]() |
#274 |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() I have installed the new TRIAC. A NTE56010. Same as before.
I added 15A fuses on MT1 and MT2 and a 0.1A fuse on GATE. I did not solder the fuses. I did a quick job. The wire is just twisted in the fuse holder's holes. This will be done properly after it's fixed. I replaced the wires from the PCB to the TRIAC so there is no splice or colour changes. It's now easier to follow. I plugged the waterbath in series with a 100w bulb and made the following measurements. First, the bulb in series drops the voltage to 77v between live and neutral. I don't know if this is normal. I also removed the heating element live wire and put a 100w bulb instead. Resistance is max between any TRIAC pins and ground with my multimeter at the max Ohm setting. Putting both multimeter leads on the heatsink also gives me a max resistance. It should give me continuity or at least some resistance. It is connected to ground. There is no continuity between heatsink and ground either. The heatsink must have a very high resistance. Why is this the case? It makes me doubt that the TRIAC is not shorted. My multimeter works fine and the leads are clean, so are the surfaces I m measurement on. There are the measurements: GATE - MT1 : 30 Ohms GATE - MT2 : 47 Ohms MT1 - MT2 : 17 Ohms GATE - Neutral : 35 Ohms Bulb (so MT2-Neutral): 12.1 Ohms Of course these measurements were done when the waterbath is off so the TRIAC is not activated. I then turned on the waterbath with the bulb in series and measured this: GATE - MT2: 77v GATE - MT1: 0.1v GATE - Neutral: 77v MT1 - MT2: 77.1v Bulb (so MT2 - Neutral): 0v At this point I haven't tried plugging the water bath staight into the wall. I think the TRIAC is OFF because it's at the same potential then live and neutral, hence why the bulb in place of the heating element does not turn on. I believe this could also be due to the fact that only 77v is supplied to the waterbath. This might not be enough to power the whole thing and turn on the TRIAC. I m not sure. I m not confident enough at this point to plug it into the wall. I have fuses on each TRIAC wire but they are 15A and 0.1A so I think the TRIAC could still blow. This is where I need your help figuring out what to do now. Anything else I can do before plugging it into the wall? Thank you! Last edited by vrasp; 07-26-2021 at 04:49 PM.. |
![]() |
![]() |
![]() |
#275 |
Solder Sloth
Join Date: Nov 2012
City & State: CO
My Country: USA
Line Voltage: 120VAC 60Hz
I'm a: Hobbyist Tech
Posts: 7,161
|
![]() Proverb: "The fast acting transistor always blows first to protect the fuse." -- unknown
|
![]() |
![]() |
![]() |
#276 |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() The reason I said it could still blow is that the max GATE current is below 0.1A so I m worried about that. Or thats what I understood from the datasheet. Also maybe I should get a 10A fuse instead. I dont know how fast a fuse blows but what if a current just over 15A has time to blow the TRIAC before the fuse blows. I have no idea if this is possible. Do you think there is a risk?
I ve gone through enough TRIACs already. Last edited by vrasp; 07-26-2021 at 06:12 PM.. |
![]() |
![]() |
![]() |
#277 |
Badcaps Veteran
Join Date: Jul 2011
City & State: Some times Sunny Jacksonville FL
My Country: USA
Line Voltage: 120 Volts 60 HZ
I'm a: Knowledge Seeker
Posts: 3,554
|
![]() I could be wrong but I would concern that it seems like you do not have a control signal to the transformer but the question is why
Because if your light is not lighting up this not making any cents to how it blowing up the triac |
![]() |
![]() |
![]() |
#278 | |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() Quote:
If there is 77v on the primary of the toroidal transformer (for the control PCB and display) then the secondary voltages must be much lower than supposed to be. That could be insufficient. The display LCD does works though. I wonder why a 100w bulb drops the voltage by that much. Its resistance must be high but why... Last edited by vrasp; 07-26-2021 at 07:09 PM.. |
|
![]() |
![]() |
![]() |
#279 | |
Senior Member
Join Date: Nov 2019
City & State: Toronto
My Country: Canada
I'm a: Knowledge Seeker
Posts: 174
|
![]() Quote:
It goes Purple - GATE Black - MT2 White - MT1 Thank you! Last edited by vrasp; 07-27-2021 at 09:01 PM.. |
|
![]() |
![]() |
![]() |
#280 |
Solder Sloth
Join Date: Nov 2012
City & State: CO
My Country: USA
Line Voltage: 120VAC 60Hz
I'm a: Hobbyist Tech
Posts: 7,161
|
![]() Well you're right it is MT1 MT2 Gate, but sorry, no additional clues why it's frying. You're on your own.
|
![]() |
![]() |
![]() |