How to correctly measure a mixed low&high current
  • edugimeno
    • May 2017
    • 568
    • Spain
    Hi there!
    Im trying to determine the consumtion rates of a module Im testing. Particularly it's a Decawave DWM1001 module, which is expected to use around nanoamps during normal state so a LiFePo4 battery should last months
    This module has peak spikes of around 100mA once a second, while the TX module is active, and then goes to sleep for the rest of the time (this is when the uA rates are expected)
    But I'm having issues to measure this current.
    1.-Using a multimeter. Obviously all I see is values jumping all the time, as this is not expected to measure such a unstable current
    2.-Using my Rigol DS1052E oscilloscope. For this I use a 0R22 shunt in series with the battery.
    When I disable the Low Power mode, I measure 4-5mV (avg, with higher spikes) which is actually the expected 15-20mA current (shunt=0.22 so current is voltage * 4.5)
    But when I enable Low Power mode, which is what I actually need to measure in detail, voltage gets down to ~1mv, which means 4-5mA, not reasonable. Then I disconnect the battery from the module and I keep reading the same average voltage!!! ~1mv.
    What is worse...Then I plug my probe tip to the Rigol ground test lead, and it keeps reading 800-1200nV...WHY??
    I tried an auto-recalibration in the Rigol Tools menu, which ended correctly, but I can't correctly measure voltages under 1mV.
    Is this normal? Is this expected from this scope.
    Anyway, this is what I have. What other options do I have to fairly-decently measure a current in the range of nA?
    I know I could take a bigger shunt, like 5 Ohm so the read voltage would be bigger and more noticeable by my scope, but in this case, the spikes (100-200mA) would drop the voltage thru the shunt resistor below the acceptable supply and it would reset or erratically fail, during this burst...
    Anyone...?? Thanks!!
  • eccerr0r
    Solder Sloth
    • Nov 2012
    • 8696
    Re: How to correctly measure a mixed low&high current
    Yes this is normal due to the limitations of the scope, in what you're trying to do.
    You'll need to design a jig for extending the dynamic range of your current measurement.
    • edugimeno
      • May 2017
      • 568
      • Spain
      Ok, and what sort of jig works fine for this. As I have no idea on what to start with
      Thank you
      • eccerr0r
        Solder Sloth
        • Nov 2012
        • 8696
        I think the easiest to understand jig is using a LNA with the 0.22 ohm resistor to bring the range into something you can accurately measure with the scope and you simply ignore the regions that go out of bounds when it hits 100mA - you measure those separately - hence improving dynamic range by measuring twice. However I think many op amps that have very low noise figures aren't cheap, cheap ones like LM741 and LM358 will not do.)
        • edugimeno
          • May 2017
          • 568
          • Spain
          Ok, I understand. I think that instead of starting from scratch designing / building / soldering... a board for this I'm groing to see if there's some pre-made module using a LNA IC that I can use for this
          Thanks a lot!
          • quickname
            Cyberpower 1500AVR UPS repair
            by quickname
            I'd like to start out by thanking this forum for providing the info to be able to repair my 15 year old CP1500AVRLCD UPS. Here is all the info I gathered during my trouble-shooting and re-capping.
            Like clockwork, I've had to replace the SLA batteries every 4 years. It was about that time and the UPS was acting like it needed new batteries... randomly shutting off with AC present, not being able to charge fully, and poor runtime when AC was removed. So, I replaced the batteries but noticed that it was not charging properly.
            I decided to verify that the batteries were...
            11-12-2023, 09:57 AM
            Is there any high current pouch battery that are size 60mmX80mm maximum thickness 21mm for a battery operated tab welder
            by sam_sam_sam
            I have a battery operated tab welder and the battery does not have enough current for the battery tab welder because after about 30 welded it does not have the same current output it voltage has gone down by at least 0.500 volts or maybe a more than that
            I can fit 21700 battery cells in the enclosure it is very tight fit for four of them that are 35 amp high current capacity 35X4=140 high current output this should work a lot better than what is installed right now
            I tested the battery cells there are installed right now and they are testing good with no issues that I...
            02-04-2025, 08:08 PM
            Diodes AP21X2SG-13 AP21X2MPG-13 Current Limited Power Switch for Laptop Computer Datasheet
            The AP2182 and AP2192 (AP21X2SG-13 AP21X2MPG-13) are integrated high-side power switches optimized for Universal Serial Bus (USB) and other hot-swap applications. The family of devices complies with USB 2.0 and available with both polarities of Enable input. They offer current and thermal limiting and short circuit protection as well as controlled rise time and under-voltage lockout functionality. A 7ms deglitch capability on the open-drain Flag output prevents false over-current reporting and does not require any external components.
            All AP21X2SG-13 AP21X2MPG-13 devices are available...
            10-23-2024, 03:33 AM
          • DynaxSC
            Extreme High Current & Low Voltage Short Killer ??
            by DynaxSC
            Hi Folks
            I'm looking for a extreme high current short killer solution for a reasonable cost.
            I have an LGA1700 motherboard (ASUS PRIME Z790-P WiFi) where I exchanged the LGA socket already 3 times, and every time after the socket exchange I get a short of power supply lines VCORE (V variable) and VCCIN_AUX (1.8V).
            After taking off the socket, the shorts dissappear again.
            The soldermask on the socket is already partially demaged from the heat, so I have reconstructed the soldermask by manually painting the missing soldermask parts with a very small top of a...
            10-15-2024, 10:13 AM
            Adjustable Boost Power Supply DC-DC Constant Voltage Current Charging Module
            by sam_sam_sam
            I want to know if I could hook up a solar panels to this boost converter would I have any issues doing and I would have issues what would they be
            https://www.ebay.com/itm/116497313155?var=416785404161
            Here are some of the description from the eBay seller
            The main parameters:
            1. Product name: High-power non-isolated DC boost module power supply.
            2. Product model: 100A2000W.
            3. Input voltage: DC12V-60V.
            4. Input current: 100A (maximum)....
            05-18-2025, 09:45 AM
